I don't have a good answer explaining this behavior yet (and have not seen but also not looked for it before). But I do have a couple of questions to better understand it:
|
Beta Was this translation helpful? Give feedback.
-
|
I am sorry for not providing enough information. Below is a comparison plot with a Rhode&Schwarz ZNLE (https://www.rohde-schwarz.com/de/produkte/messtechnik/vektornetzwerkanalysatoren/rs-znle-vektornetzwerkanalysatoren_63493-447936.html) The upper plot depicts the resonance frequency of a S21 (transmission) measurement. The devices from Rigol and Rhode&Schwarz go with their factory calibration and I calibrated the LibreVNA with a SOLT calibration at 0dB assuming ideal short, load, open, and through standards (I did not use the LibreVNA calibration kid but the standards that where in the LibreVNA package). With the exception of a slight offset (could be attributed to a different error model in the calibration or different calibration standards?) the behavior of the LibreVNA and the premium vector network analyzer from Rhode&Schwarz are similar. The phase behavior, however differs. Please note that compared to my The data for the plot was acquired as follows (I sadly don't have a touchstone file for that): The LibreVNA was operated for 30minutes before the start of the calibration and the measurement in order to properly heat it up beforehand. A heat sink was additionally used to prevent overheating. |
Beta Was this translation helpful? Give feedback.
-
Beta Was this translation helpful? Give feedback.

Thank you very much for the details. I just checked on one of my units and can see a similar behavior (about 3-4° of phase drift in S21 when changing the power from -40 dBm to 0 dBm). For this measurement I connected a short SMA cable directly between the ports:
I think this is due to different components in the LibreVNA (e.g. mixers and amplifiers) getting closer to their saturation point. In details like this it just can not keep up with professional VNAs from R&S. You should be able to improve the situation by repeating the calibration process for different amplitudes but of course that is also much more work.